I recently got a problem with my Q exective MS, the MS1 signal suddenly droped from 10E9 to 10E7, although background signal also droped, and the spectra looks normal too. The MS2 looked good too with 10E7 signal. I searched the data, and still gives me lots of proteins IDs. I switched from nanospray to regular HESI sourse, and the signal is still low. Does anyone here have a clue about this problem?
thank you!
SunSun
help! MS1 signal drop in QE MS
-
- Ubiquitin Member
- Posts: 50
- Joined: Mon Jun 03, 2013 12:33 am
-
- Angiotensin Member
- Posts: 31
- Joined: Fri Sep 16, 2011 4:41 am
Do you normally clean your ion transfer tube? If it is dirty and starting to clog, you will experience a drop of signal. Otherwise it could be dirty front optics even if your endurance test is giving good results.
Easier thing to test would be replacing the used tube for a new one and see what happens.
Easier thing to test would be replacing the used tube for a new one and see what happens.
-
- Carbon Member
- Posts: 4
- Joined: Thu Jan 23, 2014 1:17 am