Resolution vs m/z range dependency on Orbitrap

Whether you are building your own mass spec, modifying an existing one, or just need help with instrument settings post about it in here.
Suola
Proton Member
Proton Member
Posts: 1
Joined: Thu Jan 19, 2012 10:29 am

Resolution vs m/z range dependency on Orbitrap

Postby Suola » Tue Jun 27, 2017 8:33 am

Hi all mass spec enthusiasts,

In NMR, the detection scheme is in principle very similar to that of Orbitrap, i.e., Fourier transform of image current on the detection circuitry (in time domain) to human readable spectrum (in ppm or m/z domain).

Further, in NMR, the digital resolution of the spectrum depends on the desired spectral width: the narrower the spectrum, the more closely the collected datapoints are in the spectrum, and thus higher the digital resolution.

Using this analogy, and not understanding all the intricacies of Orbitrap detection, I was wondering whether similar advantage can be obtained in Orbitrap by collecting narrower spectra, say m/z range 300-1300, instead of 300-2000, in order to get higher m/z resolution in the 300-1300 range?

Best,
Sami

John F
Proton Member
Proton Member
Posts: 2
Joined: Tue Aug 19, 2014 5:46 am

Re: Resolution vs m/z range dependency on Orbitrap

Postby John F » Tue Mar 20, 2018 12:40 pm

Hi Sami,

In this case, the resolution is entirely dependent on the amount of time spent "listening" to the ions. The m/z range measured doesn't affect resolution, only the time spent "listening" to them.

You do, however, get a different advantage from a narrower scan range - improved signal. This is because the efficiency of the ion optics increases as the scan range narrows, and also because the IT settings (say 1e6 ions) is based on all the ions that are present for a scan. So, for MS1, if you have a wide range of ions and they sum to 1e6, then an ion of interest at 1000 m/z will be some percentage of the total intensity, and the absolute amount of these ions will be higher with narrower scan ranges. Practically speaking, this effect is heavily dependent upon sample complexity.


Return to “Instrumentation”

Who is online

Users browsing this forum: No registered users and 2 guests